Instruments and detector signatures¶
A spectrum on its own does not say how a dye looks to a cytometer. What the instrument sees depends on which lasers it has and which detector bands collect light. BFSpectra therefore lets you choose an instrument and converts each dye into its detector signature.
Choosing an instrument¶
The instrument box in the middle column lists every definition in
C:\Users\Public\Documents\BFSpectra\Devices\, plus None.
When an instrument is chosen:
- the Lasers list shows one tick box per laser. Untick a laser to leave its detectors out of every calculation - useful for simulating a configuration with fewer lasers;
- the plot switches to bars;
- if two or more dyes are checked, the spillover window opens alongside - see Select Colors and the spillover window.
The plot title names the instrument and the checked laser lines.
How a signature is computed¶
For each checked laser and each of its detectors, a dye's signal is
excitation efficiency of the dye at the laser wavelength x emission of the dye summed over the detector's band, in 1 nm steps
Doing this for every detector gives the signature: a vector with one number per detector. The bar chart shows one group of bars per detector, one semi-transparent bar per checked dye, in the dye's colour.
Because both spectra are normalised to a peak of 1, signatures compare dyes' spectral shape as the instrument sees it. They do not include brightness (extinction coefficient and quantum yield), laser power or detector sensitivity.
Stokes Only¶
A detector whose band starts at or below its own laser's wavelength would mostly see scattered laser light rather than fluorescence. Tick Stokes Only to leave such detectors out of the signature. This affects the bars, the spillover window and Select Colors.
Detector names¶
Each detector is labelled as
<laser>nm - <detector> <centre>nm
for example 405nm - V7 542nm. The same names are used in exported matrices and, in the
slightly different form <laser>nm - <centre>/<width>, as parameter names in
synthetic test data.
Adding an instrument¶
An instrument is a small text file. Copy an existing one in the Devices\ folder, rename
it, and edit the laser sections and detector bands - the format is described in
File formats. Restart BFSpectra to see it in
the list.