Troubleshooting¶
"Please select at least one data file!"¶
Select one or more data sets in the Data Sets panel before choosing the command.
A file does not appear after opening¶
It contained no events, or its format was not recognised. For text files, check the settings in the File Format dialog.
"PCA needs at least two parameters"¶
Tick at least two channels in the parameter list of the 2D Plot control panel.
"Void data file"¶
The data set has no events - often a gate that caught nothing.
"Select 2 or more spectrums, pls" in Decomposition¶
Tick at least two spectra - and at least two channels - in the Decomposition dialog. Save spectra first from the Average window.
"Data length and spectrum length do not match"¶
A chosen spectrum was measured with fewer channels than the data. Use spectra measured on the same instrument setup, or leave out the higher channels.
"Sorry, all selected files must have the same number of parameters"¶
Phenotyping needs files with identical channel sets.
"Number of gates differs" / "Names of the gates must be the same" in Combine¶
Every data set added to Combine needs the same gates under the same names. Use Tools → Breed gates to copy them.
The Spectra window is empty¶
The spectra folder with its names file is missing next to CytoSpec.exe.
Everything is slow¶
Set Limit Event Number in Properties, or gate out the events you need and work on the smaller data set.