Black Mass

Tick Black mass in the Elements window to analyse the spectrum as the black mass of recycled Li-ion batteries. It stays ticked for the spectra opened next, until it is unticked. The analysis is the same as for any spectrum; a black mass adds three things.

The elements of a black mass

The elements enabled in the periodic table are analysed. When none of them or all of them are enabled - the table starts with all enabled - the elements a black mass is made of are taken instead:

Element Comes from
Li Cathode
Ni, Co, Mn Cathode NMC, NCA, LCO, LMO
Fe, P Cathode LFP (Fe also from the casing, P also from the electrolyte)
Al Cathode foil
Cu Anode foil
C Anode graphite
F Binder and electrolyte - its lines are faint in a plasma in air
Na, Mg, Si, Ca, Ti, Zn Impurities

Their role

The Role column of the Elements window shows where each element comes from in a battery, as in the table above.

The cathode

The summary names the cathode chemistry the metals found point to: NMC for Ni, Mn and Co, NCA for Ni and Co without Mn, LCO for Co alone, LMO for Mn alone, LFP for Fe and P, and LFP or LMFP when Mn comes with them. Al is in every black mass, from the foil, so it can not tell NCA from NC.

P has faint lines, and its two strongest, 213.62 and 214.91 nm, sit on Cu II lines, so they are not among its key lines. When Fe is found without P, the cathode reads LFP? - the Fe may also come from the casing.