The periodic table¶
Mendeleev opens the Periodic Table of the Elements. It decides which elements - and which of their ions - LIBSight uses to explain a spectrum. The choice is remembered between sessions.
Reading the table¶
Each cell shows an element; enabled and disabled elements are drawn on different backgrounds. The panel on the left shows the selected element's symbol, name, Atomic Number and Atomic Weight.
Enabling elements¶
Right-click an element for its menu:
| Command | Effect |
|---|---|
| Enabled | Include or exclude this element from the analysis |
| Enable All | Include every element |
| Enable None | Exclude every element |
| Color | Choose the colour of this element's lines in the chart |
Every change analyses the open spectrum again straight away.
Fewer is better
Enable only the elements that can plausibly be present. The more elements are enabled, the more line coincidences there are, and the easier it is for a peak to be put down to the wrong element.
Choosing ions¶
Left-click an enabled element. Its ions appear in the list - I for the neutral atom, II
for the singly charged ion and so on - each with a tick box. Click an ion to show its lines,
with the strengths from the chosen NIST column, in the chart below the list.
Untick an ion to leave it out of the analysis. Only the neutral atom (I) and the singly
charged ion (II) are analysed - the higher ionisation states are too weak in a laser plasma - and
when both are unticked, both are used. The ticks also choose the ions of the element's
overlay on the main chart, below. The tick box above the list ticks or unticks all ions of
the element.
Element spectra on the main chart¶
Left-clicking an element also switches its overlay on the main chart on or off: the model spectrum of its ticked ions, in the element's colour, scaled to the measured peaks. This is the quickest way to see whether an element's lines line up with the spectrum. Click the element again to remove the overlay.
Where the lines come from¶
The line lists are taken from the NIST Atomic Spectra Database and stored in the elements
folder, one file per element, with the lines from 200 to 1999 nm. Before version 1.3 they
stopped at 632 nm; Li I 670.78 and K I 766.49 and 769.90 are among the lines added since.
Only well-measured lines are kept: lines observed in absorption,
extrapolated, tentatively classified, masked, of uncertain identification, roughly
estimated, hazy, diffuse, shaded or forbidden are left out. Which NIST column is used as the
line strength, the line width, and the accuracy limit are set in the
options.