Options¶
Options opens the Settings window. Apply stores the values and analyses the open spectrum again; Default fills in the default values (press Apply to use them); Close hides the window.
| Setting | Meaning | Default |
|---|---|---|
| NIST DB Column | Which NIST value is used as the line strength: Intensity (relative observed intensity), Probability (transition probability, Aki) or Strength (line strength) | Intensity |
| Minimal WL | First WL of the First Spectrometer - start of the analysis grid, nm | The first line of the database, 200 nm |
| Maximal WL | Last WL of the second spectrometer - end of the analysis grid, nm | The last line of the database, 1999 nm |
| Merge Shift | Inter spectrometers shift - applied to the second spectrometer's spectrum before merging, nm | 0.18 |
| Total Shift | Shift of the combined spectra - a wavelength calibration correction, nm | 0 |
| Spectrum Size | Number of points in combined spectrum - resolution of the analysis grid. Opening a spectrum sets it to the number of points of that spectrum | 8000, until a spectrum is opened |
| Line St Deviation | Standard Deviation of WL for Pseudo Spectra - the width of every model line, nm. A key line is found when a peak is within twice this of it, or within 1.5 points of the spectrum where those are further apart | 0.075 |
| Max Error | Maximal Deviation of Line Intensity - lines whose NIST accuracy is worse than this are left out; 3 accepts every line | 3 |
NIST rates the accuracy of transition probabilities with letter classes; LIBSight converts them to fractional uncertainties:
| Class | AAA | AA | A+ | A | B+ | B | C+ | C | D+ | D | E |
|---|---|---|---|---|---|---|---|---|---|---|---|
| Uncertainty | 0.003 | 0.01 | 0.02 | 0.03 | 0.07 | 0.10 | 0.18 | 0.25 | 0.40 | 0.50 | 1.0 |
Lines without a class count as 2.
Note
Opening a spectrum sets Minimal WL and Maximal WL to its own wavelength range, and Spectrum Size to its number of points. To analyse part of the spectrum only, or on another grid, change them and press Apply; the next spectrum opened resets them.
Choosing the line width¶
Line St Deviation should roughly match the width of the lines your spectrometer produces. Too narrow, and small calibration errors make key lines miss their peaks; too wide, and the peaks of other elements' lines nearby count as found.