Troubleshooting

"NO DATABASE FOUND" at start-up

The elements folder is missing from the program folder. Reinstall LIBSight.

"spectrometer" not found. Using: ...

The file name contains neither spectrometer_1 nor spectrometer_2, and the file was not recognised as a blended table, so it is opened alone as spectrometer 1. Rename the pair, or save the spectrum as a blended .csv.

"No spectrum found in ..."

The blended .csv file has fewer than two usable lines. Check that the first column is a wavelength and that the numbers use a dot as the decimal separator.

Peaks without labels

No element found explains them. The summary of the Elements window lists the strongest of them with the elements that have a strong line there; or move the cursor onto the peak, widen Tolerance if needed, and look at the ion table for candidates. Enable those elements in the periodic table.

An unlikely element is found

Look at its Lines and Key line in the Elements window, and click its row to draw the peaks put down to it on the chart: a line of an element that is present may sit on one of its key lines. Disable the unlikely element, or untick its ions, and compare. An element found with one key line only says so in the Note column.

The two spectrometers do not join smoothly

Adjust Merge Shift in the options until lines in the overlap region coincide.

Unmixer II refuses to run

  • Load a medium/solvent spectrum first - choose the medium file.
  • These are not on the wavelengths of the medium - the listed files were measured with a different range or number of points. Use files from the same instrument set-up.

Resetting LIBSight

Close LIBSight and delete %APPDATA%\LIBSight\LIBSight.ini.