Unmixer II¶
Unmix opens Unmixer II. It works without the element database: it explains a sample spectrum as a mixture of spectra you measured yourself - the solvent and a set of controls, each a known substance measured on the same instrument.
Setting up¶
| Field | Content |
|---|---|
| Background/Dark Current File | A spectrum taken with the laser off - the detector background. Select chooses it |
| Medium/Solvent File | A spectrum of the solvent or carrier alone. Select chooses it |
| Controls | The control spectra. + (Add Control) adds one or several files; - (Remove Control) removes the ticked controls, or the selected one if none is ticked |
| Sample File | The spectrum to analyse. Sample chooses it |
Any of the spectrum file types can be used, including .sif series.
The chosen files are remembered for next time.
Click a file field, or a control in the list, to show that spectrum in the chart.
Running¶
Press Run. Unmixer II:
- subtracts the background from the medium, every control and the sample;
- subtracts the medium from every control, so each control contributes only its own lines;
- fits the sample with the medium and the controls - for a
.sifseries, shot by shot; - shows the BarChart with one bar per component - the medium first, then each control, labelled with the first two letters of its file name - giving the mean amount and its standard deviation;
- draws the Sample and the Composition - the medium and controls at their fitted amounts - on the chart.
Same wavelengths
Unmixing compares the spectra point by point, so every file must have the same number of points over the same wavelength range as the medium. If one does not, Run stops and lists the files that differ. A medium must be loaded before running.