Clusters and PCA¶
Xploid groups wells with similar DNA histograms, which quickly shows which samples on a plate belong together - diploids, tetraploids, mixoploids, failed samples.
Clustering¶
The number of groups is set with Cluster Amount on the Plate Map tab. How similarity is measured is set on the Options tab:
| Setting | Choices |
|---|---|
| Metric - distance between two histograms | Euclidean, Euclidean Relative, Earth Mover, Kolmogorov-Smirnov, Kullback-Leibler |
| Intercluster Distance - distance between two groups | Centroid, Min, Max, Mean, Ward |
Earth Mover and Kolmogorov-Smirnov compare the shape of the histograms and are least affected by small shifts in peak position.
The Clusters tab shows one row per group with the group's histogram. Click a row to select all wells of that group.
The plate map¶
The Plate Map tab shows the plate in one of two ways:
- Show Heat Map ticked - every well is a circle coloured by the fraction (Percent) or number (Count) of its events that pass the noise gate, as chosen in Heat Map Values. The colour scale is shown beside it.
- Show Heat Map unticked - every well shows its histogram, coloured by its cluster. Hide Clusters draws them all in grey.
| Control | Effect |
|---|---|
| Keep Aspect | Keep wells round when the panel is resized |
| Copy | Copy the plate map to the clipboard |
| Set Wells for Total | Use the selected wells for the total histogram - see Opening plates |
Click a well to select it, drag to select several. Point at a well to see its values and file name in the status bar; the status bar also shows the plate total.
PCA¶
The Plot tab projects every well's histogram onto its first two principal components. Wells with similar histograms end up close together.
| Control | Effect |
|---|---|
| PCA | Compute the projection |
| Number box | How many bins each histogram is reduced to first - 256 down to 2 |
| PCA live | Recompute whenever the clustering changes |
Double-click the chart for the chart editor.