Hybrids and aneuploidy¶
When every sample is co-stained with an internal standard - nuclei of a reference with known DNA content - the ratio of the sample's peak to the standard's peak measures DNA content independently of instrument drift. The Hybrid tab turns that ratio into a C-value, a ploidy level and a chromosome count.
Setting up¶
- Define a peak for the internal standard and a peak for the samples' G1 - see DNA peaks.
- In the well table, right-click the wells that are controls (samples of known ploidy) and tick Control. Wells that hold only the standard can be ticked Internal Standard.
- On the Hybrid tab choose:
| Setting | Meaning |
|---|---|
| Mode | Interspecific Hybrids or Aneuploidy Test - see below |
| Internal Standard Peak | The peak of the internal standard |
| Control Standard Peak | The peak of the samples, as measured in the control wells |
| C-Value | The DNA content of the standard, to express results as C-values |
| Ploidy Level | The ploidy of the controls |
| Haploid Set | The number of chromosomes in one haploid set, for chromosome counts |
| P-Value | The significance level for detecting deviating samples; the matching Z-Value is shown |
| Use Median instead of Mean | Use the median ratio as the reference |
Apply recalculates. The status line reports what is missing, for example Set Internal Standard please.
If no control wells are marked, Xploid tries to find the controls itself: the largest group of samples with a consistent ratio, within the chosen P-value.
The calculation¶
For every sample well:
| Result | Calculation |
|---|---|
| Ratio | R |
| CV | How far R is from the mean (or median) ratio, in percent |
| C-Value | R × C-Value |
| Ploidy | R scaled so that the controls come out at Ploidy Level |
| Chromosomes | Ploidy × Haploid Set, rounded to a whole number |
Interspecific Hybrids scales by the ratio of the control peak to the internal standard. Aneuploidy Test scales by the mean ratio of the control wells, so small departures from the controls stand out.
A chromosome count that is not a multiple of the haploid set is written with the difference,
for example 18+1 or 36-2, and highlighted.
The label under the settings shows the mean (or median) ratio and the coefficient of variation of all ratios.
Results¶
- The chart on the Hybrid tab plots, for every sample, its ploidy level - or its C-value if no ploidy level is given, or its plain ratio.
- The Ploidy Table tab lists Sample, G1 Std (standard peak position), G1 Hybrid (sample peak position), Ratio, CV, C-Value, Ploidy and Chromosomes.
- Export on the Hybrid tab writes the Ploidy Table as CSV.
- The same results appear in the report.