DNA peaks¶
A peak is a band on the DNA histogram - G1 of the diploid population, G2 of the diploid population, G1 of a tetraploid, the internal standard, and so on. Xploid fits a curve to the histogram of every well inside each band and reports where the peak is and how big it is.
The histogram¶
The histogram chart shows the reporter histogram of the selected wells, or of the total wells when nothing is selected.
| Control | Effect |
|---|---|
| REPORTER | The DNA parameter. Double-click or right-click it to make it the default reporter |
| X axis | Click to show another parameter. Peaks are shown only on the reporter |
| Selected Wells | Show only the wells selected on the plate map or in the well table |
| Auto Scale | Rescale the axes to the data |
| Separate Wells | Draw each selected well in its own colour instead of summing them |
| Show Gates | Show the peak bands |
| Show Peaks | Show the fitted curves |
Double-click the chart for the chart editor.
Adding peaks¶
| Button | Adds |
|---|---|
| + | A peak placed on the next peak of the total histogram to the right of the existing peaks, with the shape chosen in Peak Shape on the Options tab |
| +S | A mixture peak - several overlapping components fitted together, for a broad or shouldered peak |
| +A | All peaks at once, found automatically. Existing peaks are removed first; you are asked before that happens |
New peaks are named with the Default Peak Name Prefix and a number - G1, G2 ...
Adjusting and naming peaks¶
Drag a band's edges or the band itself to change the range it covers. Click a band to open the Peak Name dialog:
| Control | Effect |
|---|---|
| Pick a name | The peak's name, used in tables and exports |
| Pick Type | Gaussian, Log Gaussian or Gaussian Mixture |
| Delete Gate | Remove the peak |
Default diploid peak
If Default Diploid Peak Number on the Options tab is set, the peak whose name
carries that number - G2 for 2 - is used as the diploid peak for
ploidy ranges automatically.
Peaks in individual wells¶
In the well table, every well's histogram shows a small flag on each fitted peak. Click a flag to:
- exclude a peak that was fitted but is wrong in this well;
- include it again;
- accept a peak that was below the noise level and therefore ignored.
The fit, ploidy and results update immediately.
Shifting one well¶
Instruments drift. If one well's histogram is displaced relative to the bands, select that well alone and move the Shift slider. The well's histogram is shifted by that many channels and refitted.
Fitting settings¶
On the Options tab:
| Setting | Effect |
|---|---|
| Peak Shape | Normal or Log-Normal for new peaks |
| Minimization Algorithm | Chi Square, Marquardt, Simplex or BFGS |
| Noise Level (in standard deviations) | Peaks weaker than this, relative to the background, are ignored |
| Use Base Line Compensation | Fit and subtract a polynomial background under the peaks - useful with a lot of debris |
| Base Line Polynom Order | The order of that polynomial |
| Size of Flags on Histos | The size of the peak flags in the well table |
Markers¶
Markers are vertical lines at fixed positions on the reporter axis - for example, where 2C and 4C of a reference line usually fall - drawn on the histogram and the gate charts.
On the Options tab, + adds a marker (enter its position as a number times a power of ten), - removes the selected one, and the tick box next to Markers shows or hides them. A marker can also be dragged on the chart. Markers are kept between sessions.