Ploidy ranges

Xploid assigns every well a ploidy level by looking at where its relevant peak falls on the DNA axis. The ranges for each level are set on the Ranges For Ploidy tab.

How the ranges are built

Everything follows from one value, the position of 2C - the G1 peak of a diploid:

  • Select Diploid Peak - use the position of one of your peaks; or
  • or enter Reference 2C - type the position yourself.

For each level N of 1, 2, 3, 4, 6, 8, 12, 16, 24, 32 and 64, the centre of the range is

\[ c_N = N \times \frac{\text{2C}}{2} \]

The upper bound is cN × (1 + p), where p is Range (%). The lower bound is the upper bound of the level below, so the ranges join up; the lowest one starts at c1 × (1 - p).

The table on the tab shows, for every level (C1, C2, C3 ...), its Mean (centre), Low and High bound.

Which peak counts

Small peaks are ignored so that a few stray nuclei cannot decide a well's ploidy:

Setting Meaning
Minimum (%) A peak must hold at least this percentage of the count of the largest peak
Minimum (abs) The same threshold as an absolute count, shown for the current plate

Using the peaks instead

Tick Use Peaks instead of The Table to take the ranges from your peak bands rather than from the reference: each band is one level.

Or press Create Peaks from The Table to do the opposite: replace the peaks with one band per level, named C1, C2 ... and fit them. This is a quick way to set up a plate of unknown samples.

Results

  • The Ploidy column of the well table shows each well's level on its colour.
  • The Ploidy Levels tab shows the same numbers laid out as the plate.
  • The levels are part of every export and of the report.

A well with no qualifying peak gets level 0.