Skip to content

Scanning planned plates

Planned plates are scanned with BacBuster, on a BARDOT instrument. BacBuster has ExpManager built in, on its Project tab, and takes each plate's description from the database - so the person at the instrument does not type it again.

In BacBuster, by hand

On BacBuster's Project tab, choose the user and the experiment, then:

  • drag a plate from the list onto BacBuster's plate image, or
  • scan the dish's label with a barcode scanner.

BacBuster takes the plate's name, barcode and description from the database, and the plate turns yellow in the list. Scan it as usual; when BacBuster saves it to the database, the scan replaces the planned plate. The BacBuster manual describes it step by step.

In BacBuster, with a Cytomat

The Cytomat window of BacBuster lists an experiment's planned plates for each of its places, by name. A Cytomat run then scans the plates at the times set, and saves each under its name - see the BacBuster manual.

Barcode scanners

ExpManager and BacBuster read a barcode scanner that types like a keyboard. So that a scan is not mistaken for typing, the scanner must send:

Before the code The code After the code
STX, Ctrl+B (ASCII 2) The barcode's characters ETX, Ctrl+C (ASCII 3)

Set the scanner's prefix and suffix that way - its manual shows the codes to scan to do so. The whole code has to arrive within 5 seconds.

In ExpManager, scanning a label highlights the plate in yellow. If the plate belongs to another user or experiment, ExpManager switches to it first.

Following progress

The plate list shows planned and scanned plates alike. To see which are done, open the experiment in Baclan or in BacBuster's Database Control Panel - scanned plates have an image.