The Cytomat incubator

With a Thermo Cytomat automatic incubator beside it, BacBuster scans plates by itself: at the times you set, the Cytomat hands a plate out, BacBuster scans it and saves it, and the plate goes back to incubate - for as long as the experiment lasts, around the clock.

Turning it on

Tick Cytomat on the Service > Staff Only tab and restart BacBuster. The Cytomat button appears on the toolbar; it opens the Cytomat window. The Cytomat's serial port is chosen as Port on the window's Advanced tab - BacBuster opens it at start-up.

The Cytomat window

Places

The Cytomat holds two towers of 21 places, on the tabs Tower 1 (places 1-21) and Tower 2 (places 22-42); place 1 is at the bottom. Each place is one row:

Column Meaning
## Tick to use the place. Unticked places are left alone
Experiment The experiment the plate belongs to; the list comes from the database the window names
Plate The plate's name; the list holds the experiment's planned plates. Double-click it for the plate's details: user, organism, lab, medium, source
Plating Date, Plating Time When the dish was plated; the incubation time saved with each scan counts from here
Start Date, Start Time The first scan
D(h) Duration in hours: no scans after start + duration
S(m) Step in minutes between scans. 0 scans once
^ Copies the row to the next place - see below

A place is scanned at the start time, then every step, until the duration is up. The row's colour shows its state, and Next Run at the top shows when the next scan of any place is due, with the time left.

Filling many places

The ^ button of a row copies it to the next ticked place. The list beside CheckAll chooses the direction:

Choice Effect
^ Copy Up Copies to the next ticked place above
^^ Shift Up Shifts all the rows above one place up, starting with this one
v Copy Down Copies to the next ticked place below
vv Shift Down Shifts all the rows below one place down

Holding Shift, Ctrl or both while clicking picks Shift Up, Copy Down or Shift Down. The column headings choose what is copied: click a heading to copy only that column, click more headings to add them; when all are chosen, everything is copied. CheckAll and CheckNone tick and untick all places.

Running

Button Does
Start Starts the schedule; the button becomes Stop. While it runs, most of the window cannot be edited
Run Right NOW Scans every ticked place once, now, then carries on with the schedule
Save, Load Save the schedule to a file, and load one

Before starting, BacBuster checks the image options: Colorify, Raw Plate Image and Positive Image should be off for a run, and BacBuster offers to turn them off.

For each scan, BacBuster moves its stage to the loading position, the Cytomat puts the plate on it, and BacBuster takes the plate image, scans the colonies and saves the plate to the database under the place's experiment and name, with the user Cytomat and the incubation time since plating. Then the Cytomat takes the plate back. When several places are due, the one most overdue goes first.

The schedule is kept as it is changed. If BacBuster is closed during a run and started again, the window opens and the run carries on.

Plate out

While a plate is out of the Cytomat, no other is fetched. If a run was interrupted with a plate out, take it back by hand, then click the Plate Out button and confirm Click yes if Bardot device has no plate inside.

Status

The lights at the top turn red when:

Light Means
Busy The Cytomat is moving
Ready The Cytomat reports ready
Warning, Error The Cytomat reports a warning or an error. An error also brings up the red warning button on BacBuster's toolbar and sends the e-mail alert
Shovel occupied A plate is on the Cytomat's shovel
Gate Open, Door open The Cytomat's gate or door is open
Table occupied A plate is on the transfer table

The box beside the lights logs what the Cytomat was told and answered.

Temperature

Temperature shows the Cytomat's temperature. With Remote SetPoint ticked, BacBuster keeps the Cytomat warm with the instrument's heater:

Field Meaning
Set point The temperature to keep, °C
Cycle, s The heater coil's cycle, 10 to 3600 s
Duty, % The part of each cycle the coil is on, 25 to 100 %

Below the set point the fan runs and the coil switches on and off in cycles; at the set point the coil stops, and the fan runs on for a few minutes to carry the heat away. The coil never runs without the fan. The settings are kept across restarts.

E-mail alerts

The Options tab has a box of e-mail addresses, one per line. When a run cannot save a plate or the Cytomat reports an error, BacBuster sends a message to each. Test Emails sends a test message.

The Advanced tab

For setting up and fault finding: the serial Port and Open; initialising the Cytomat and its motors; opening and closing the gate; moving the shovel to the table, back, or to the wait position; reading the status, errors and warnings and resetting errors; and Command / Send for commands typed by hand. Tick No Table Sensor if the table's sensor does not work.