Scanning colonies¶
Scan Colonies on the toolbar takes a scatter image of every red colony on the plate image.
What happens¶
- If the plate already has scatter images that were not saved, BacBuster asks whether to save them first; scanning starts the plate's scatter images afresh.
- The laser comes on.
- BacBuster plans the shortest route through the chosen colonies - drawn as a line on the plate image - and leaves out any the stage cannot reach.
- For each colony in turn the stage brings it into the beam, centres it, sets the exposure and takes its scatter image. The image appears on the left as a thumbnail, and in the scatter image panel.
- When all are done the laser goes off, each image is graded, and the thumbnails are ticked or unticked by grade - see Reviewing scatter images.
While it works the button shows a stop sign: click it to stop after the current colony. The progress bar at the bottom left shows how far it has got, and the Log tab says how many seconds each colony took.
Centring¶
A colony is centred when its scatter pattern sits in the middle of the scatter camera. After each move BacBuster finds the centre of the pattern and moves the stage to correct it, up to Max Steps times (on the Options tab, normally 25). Which pixels count as part of the pattern is set by Finding Center Threshold for LASER Camera, low and high, on the same tab; Show Threshold shows only that range on the live scatter view, so that you can see what the centring sees.
| Option (Options tab) | Effect |
|---|---|
| Max Steps | Most centring moves per colony. Fewer is faster; too few leaves colonies off-centre |
| Finding Center Threshold for LASER Camera | The brightness range, 0 to 255, used to find the pattern's centre |
| Use Laser Offset | Takes the offset between the beam and the camera's centre, measured by calibration, into account |
| Centering Mode | Fixed to Bacteria, for colonies; calibration switches to Dots by itself |
Exposure¶
With Auto Exposure ticked on the Control tab, the exposure is adjusted for every colony until the image reaches the Magic Number brightness. Without it, every colony gets the exposure of the slider under the live scatter view. See Exposure and dynamic range.
Following colonies over time¶
When the same plate is scanned again and again - by the Cytomat or a Test Drive - the cache remembers where its colonies were, so that each run scans the same colonies as they grow, instead of whatever is found that time. Tick UseCache on the Service > Cache tab to use it, and Clear Cache on Cytomat Start to start every Cytomat run afresh. Clear Cache forgets everything; Export Cache writes the colonies' histories as text.
Mini BARDOT¶
On the rotating Mini BARDOT instruments, colonies too close to the centre of rotation cannot be reached by the beam and are left out.
After scanning¶
- The plate image shows the route and a label on the selected colony, for example
3/17- scatter image 3, colony 17. - The Colony Table tab lists the scanned colonies with their position, diameter, roundness and the stage position they were scanned at. Right-click it to export it.
- Warning! Exposure times are variable!!! Full rescan highly recommended means that, without auto exposure, the exposure was changed during the scan, so the images are not comparable.