Calibration

Calibration teaches BacBuster where things are: how many plate-camera pixels make a millimetre, how a point on the plate image maps to a stage position under the laser, where the laser falls on the scatter camera, and how bright the plate light makes an empty dish. Calibrate a new instrument, and again after anything in it was moved, adjusted or replaced. Its results are written to the instrument's .ini, .tab, .bgr and .cpt files.

Big BARDOT

The Service > Big Bardot tab calibrates the X-Y stage instruments. You need the calibration plate: a dish-sized plate with a printed pattern of dots.

  1. Set Dot diameter in tenth of millimeter to the diameter of the plate's dots, and laser beam diameter (mm) to the beam's diameter.
  2. Click Start Calibration. The stage comes out to the loading position.
  3. Put the calibration plate in and click OK on Insert Calibration Plate and click OK.
  4. BacBuster sets the cameras' exposures and records the plate camera's background, finds the laser's centre on the scatter camera, finds the dots on the plate image, and visits them with the laser to map plate-image positions to stage positions. A small window with a stop button lets you abort it.
  5. Calibration is Done, or Calibration Failed. Error: ... with the reason.
  6. The stage comes out again: REMOVE THE CALIBRATION PLATE NOW!!! Take it out.

A calibration that fails writes nothing: the previous results stay in the instrument's files. The reasons it gives:

Message Meaning
Can not find any dot!, Can not find second dot!, Can not find 3rd dot! The dots of the calibration plate were not found on the plate image: plate missing, upside down, or the exposure far off
Can not find the circle!, Can not find the circle number!, Can not find ring The plate's reference marks were not found
Calibration Aborted You stopped it

Other controls on the tab:

Control Does
Set Plate Image Postion Makes the stage's current position the one plate images are taken at
Set Laser Position Makes the stage's current position the laser's reference position
Grip, Drop Close and open the plate holder's gripper
Test Find Colonies Plate Image loads a picture from a file instead of taking one, to test colony finding
Replace Laser camera with Microscope one Scans with the microscope camera

Mini BARDOT

The Mini Bardot tab calibrates the rotating instruments. Put the calibration plate in and click Calibrate MINI; when it is done the tab shows the Rotation Center, the Deviation and the Calibration Plate Angle it found. Set Camera Exposure Times as Default makes the current exposures of both cameras the defaults. Raw Coordinate Mode moves the stage in motor coordinates, without the calibration.

Colour plate camera

With a colour plate camera, the Color Camera tab sets its white balance: the Red, Green and Blue sliders, or Auto to balance it on what it sees.

Agar thickness

A dish's agar bends the light, so the plate image's scale depends on how thick the agar is. The Agar Thickness box on the Control tab corrects for it:

Control Does
Thickness Compensation Applies the correction
Agar Thickness The agar's thickness, mm
Distance from Agar to Lens(mm) The distance from the agar to the plate camera's lens
Calculate From Current Plate Works out the thickness from where the colonies of the plate just scanned were actually found
The list, Save, Delete Named thicknesses, for the dishes you use: type a name and Save

Typing a new thickness switches the compensation off, in red, until you tick it again.

Printed plates

The Staff Only tab prints the patterns used for calibration and testing, on a printer that prints true to scale: Print Distortion Map, Print Fine Calibration Plate, Print Round Calibration Plate, Print Rect Calibration Plate and Print Round Test Plate. The Instrument is not calibrated means the distortion map needs a calibration first.