Exposure and dynamic range¶
A scatter pattern is bright in the middle and faint at the edges. How long the scatter camera exposes, and whether it combines several exposures, decides how much of the pattern is recorded.
The exposure sliders¶
The slider under each live view sets that camera's exposure. The scale is logarithmic: each step is the same ratio, not the same number of milliseconds. Hover over the plate camera's slider to see its exposure in seconds. Double-click a live view to return its camera to the default exposure, which calibration sets - or Set Camera Exposure Times as Default on the Service > Mini Bardot tab.
Auto exposure¶
With Auto Exposure ticked on the Control tab, BacBuster sets the scatter camera's exposure for every colony separately: it lengthens or shortens the exposure until the image's brightness, measured the way the instrument is set up to measure it, reaches the Magic Number (1 to 254). A second slider under the live scatter view sets the Magic Number too.
Without auto exposure, all colonies of a plate are taken with the same exposure - the one the slider sets - which keeps their images comparable. Change it during a scan and BacBuster warns that the exposures vary.
| Setting | Where | Meaning |
|---|---|---|
| Auto Exposure | Control | Exposure per colony |
| Magic Number | Control | Target brightness for auto exposure; higher gives brighter images |
| AutoExposure Alg | Service > Staff Only | How brightness is measured: from the histogram, the median, the mean square radius, the density of scattered light, or averages inside or outside the beam's centre |
| Scatter Fraction for Density or Radial decay for Average | Service > Staff Only | The parameter of the measures that need one |
The algorithm and its parameter belong to the instrument; leave them as calibration and service set them.
Dynamic range¶
Dynamic Range on the Control tab chooses what is taken of each colony:
| Choice | What is taken |
|---|---|
| Low | One image at one exposure: the normal scatter image |
| High | A high dynamic range (HDR) image instead: a series of exposures combined |
| Both | The normal image and the HDR image |
An HDR image starts at HDR Exposure (seconds) and halves the exposure picture by picture until nothing is overexposed, then combines the pictures. It is shown on a logarithmic scale spanning HDR Decades powers of ten. It records both the bright centre and the faint edge of the pattern, at the cost of time.
With Both, the HDR images are kept apart: saving to the database writes a second plate named
<plate name>_HDR, and saving to a file a second file <file name>_HDR.plt. The LDR and
HDR tabs of the scatter image panel show the two.
Zoom¶
1X, 2X and 4X under the live scatter view enlarge the centre of the scatter camera's picture. Calibration measures the scatter camera at each zoom; use the zoom your experiments are set up for, and the same one for all of them.
How scatter images are scaled¶
Intensity Map to 0..255 for Scatter on the Service > Staff Only tab chooses how the camera's brightness is mapped to the 256 grey levels of a scatter image: Linear, Hyperbolic sine or Logarithm. The last two show more of the faint parts.