The tabs on the right¶
Above the tabs, >> hides them to give the images more room, and Profile recalls stored settings - see Profiles. BacBuster reopens on the tab it was closed on.
Control¶

| Control | Does |
|---|---|
| Snap | Scans the selected colony where the stage is now and adds it to the plate |
| ReSnap | Replaces the selected scatter image with a new one |
| Delete Samples | Removes all scatter images, keeping the plate image |
| Plate Width, Plate Height, Radius, Round Plate | The area of the dish searched for colonies |
| Auto Exposure, Magic Number | Exposure set per colony, to this brightness |
| Max Colony Number | Most colonies scanned per plate; 0 for all |
| Plate image reps, Gain | Pictures added into one plate image, and the contrast curve of the sum |
| Dynamic Range: Low, High, Both | Normal scatter images, HDR images, or both |
| HDR Exposure, HDR Decades | The longest exposure of an HDR image, and the range it shows |
| My Goodness Level | Colonies graded below this are unticked |
| Reset Stage | Sends the stage home and then to its parking position |
| Agar Thickness box | Corrects the plate image's scale for the agar's thickness |
| The chart, and ShowPlot | The profile of the selected colony |
See Scanning colonies, Exposure and dynamic range and Calibration.
Find Colonies¶
The threshold, diameters and roundness colonies are found with, and the list of the colonies to be scanned - see Finding colonies.
Colony Table¶
The scanned colonies: number, position, diameter, roundness and stage position. Right-click to export it - see Reviewing scatter images.
Options¶
| Control | Does |
|---|---|
| Database Options | Opens the database Manager - see Connecting to a database |
| Colorify | Scatter images in false colour |
| Raw Plate Image, Take Raw Image, Positive Image | Plate image options - see The plate image |
| Use Laser Offset | Corrects for the offset between the beam and the scatter camera's centre |
| Speed | The motors' speed, as a pulse rate |
| Finding Center Threshold for LASER Camera, Show Threshold, Max Steps | How colonies are centred in the beam - see Centring |
| Centering Mode | Fixed to Bacteria |
| E-mail box, Test Emails | Who is told when an unattended run fails - see E-mail alerts |
Service¶
Calibration, the camera's white balance, test runs, the cache and the instrument's settings - see The Service tab.
Project¶
The plates planned in ExpManager - see Planned plates.
Log¶
Every message, with its time - see The Log tab.
Jukebox¶
The motors play a tune - see The Jukebox tab.