The Service tab

The Service tab is for setting up and maintaining the instrument. Its own tabs:

Tab For
Big Bardot Calibrating the X-Y stage instruments - see Calibration
Mini Bardot Calibrating the rotating instruments - see Calibration
Color Camera White balance of a colour plate camera
Test Drive Scanning one dish at intervals - see Test Drive
Staff Only The instrument's settings, below
Cache Following colonies from scan to scan - see Scanning colonies

Only the tab of the instrument's kind is shown: Big Bardot or Mini Bardot.

Warning

The settings on these tabs belong to the instrument and its calibration. Change them only if you know why, and note what they were.

Staff Only

The instrument

Control Does
Change Device Chooses the device configuration - see Devices and cameras. Restart afterwards
Swap Cameras, Camera Settings Which camera is which, and their orientation - see Devices and cameras
Motor Configuration Hettich instruments: the motors' settings
Ports The serial ports of the relay, motors, heater and Cytomat - see Devices and cameras
Cytomat An incubator is attached: shows the Cytomat button. Restart afterwards
NoGizmo Runs without an instrument after the next start
100% motor power, External Laser See Lights and safety
LEDsaver, LaserSaver Minutes before the plate light and the laser switch themselves off
Park, Eclipse, Fan, Coil Buttons for the stage and the heater

Images

Control Does
No Distortion Correction Plate images are not corrected with the calibration's distortion map
No Background Correction Plate images are not corrected with the calibration's background
Show Histogram Shows the histogram of the live scatter image
AutoExposure Alg, Scatter Fraction for Density or Radial decay for Average How auto exposure measures brightness - see Exposure
Intensity Map to 0..255 for Scatter How scatter images are scaled - see Exposure
Test Filter Opens the filter test window on the plate camera's picture, to try the filters colony finding could use
Save Plate Image, Load Plate Image Save the plate camera's picture to a file, and load one as the plate image

Measurements

Control Shows
x=, y=, I= The plate image pixel under the mouse and its brightness
X =, Y = The stage position the live plate view's scroll bars point at
The table of Current, Best, Laser, Salesman ... Figures of the last colony's centring and of the scan route, for fault finding
Grooves/mm, Order The diffraction grating used to measure the Distance to Camera

Printing

Print Distortion Map, Print Fine Calibration Plate, Print Round Calibration Plate, Print Rect Calibration Plate and Print Round Test Plate - see Calibration. DPI shows the printer's resolution after a print.

Recording the instrument

Use Camera, with Camera App Path and Output Directory, starts a separate camera program while a Cytomat run fetches and scans a plate, to film the instrument at work.

The Log tab

Every message BacBuster writes - plates started and saved, colonies graded, the Cytomat's commands, errors - with the time. The same messages go to a log file per day in the log folder of the program folder. See Files and folders.

The Jukebox tab

Play makes the stage motors play the tune chosen under Song - from Jingle Bells to Hail Purdue. Click Play again to stop. It does not play while colonies are scanned, nor on Hettich instruments.