Choosing colonies

Every colony that shows red on the plate image will be scanned; green ones will not. Colony finding decides the colours first - see Finding colonies - and the plate tools above the plate image change them.

The plate tools

The lasso

Pick the Lasso tool and drag around the colonies you want, as a closed loop - BacBuster closes it for you when you let go. Only colonies inside a lasso stay red; the rest turn green. Draw more lassos to add more areas; each one stays drawn on the plate image.

The crossed-out lasso removes all lassos and makes every colony found red again.

The hand

Pick the Hand tool and click a colony to switch it between red and green. Use it to take out a colony touching the dish edge, or to add back one that roundness left out.

How many

Max Colony Number on the Control tab caps the number of colonies scanned per plate; 0 means no limit. With a cap, BacBuster scans that many of the red colonies.

Scanning more of the same plate

To add one colony to a plate already scanned, select it with the Arrow tool, bring it into the beam with the Cross tool (Shift +click, see below), and click Snap on the Control tab. BacBuster centres it, takes its scatter image and adds it to the plate.

ReSnap replaces the scatter image of the thumbnail you selected with a new one, taken where the laser is now. Select Colony to replace means no thumbnail was selected.

Delete Samples removes all scatter images and keeps the plate image and its colonies, ready to scan again.

The cross

With the Cross tool, a click on the plate image moves the stage so that the laser falls on the point you clicked; Shift +click moves it to the centre of the selected colony. With the Cross tool down, clicking a thumbnail moves the stage back to where that colony was scanned and sets the scatter camera's exposure to the one it was scanned with. The live scatter view then shows that colony, ready for ReSnap.

The arrow

The Arrow tool, the normal one, only selects: a click selects the nearest colony, shows its scatter image if it has one, and plots its profile on the Control tab.