Scanning your first plate¶
This walks through one plate by hand, from putting the dish in to saving it. Each step has a page of its own with the details.
1. Check where the plate will go¶
The database label on the toolbar names the database the plate will be saved to. If it says Click to select database, click it and choose one - or plan to save the plate to a file instead. See Connecting to a database.
If the plate was planned in ExpManager, drag its name from the Project tab onto the plate image now: BacBuster fills in its name, barcode, organism and the rest. A barcode scanner does the same when you scan the dish's label. See Planned plates.
2. Put the dish in¶
Click Load / Eject. The stage comes out to the loading position. Put the dish in the holder.
3. Take the plate image¶
Click Plate Image. The stage moves the dish under the plate camera, the plate light comes on, the camera takes the picture and the light goes off again. The picture appears below the live views with the colonies found on it in red.
If too many or too few colonies are found - specks of dust, merged colonies, faint colonies missed - adjust the settings on the Find Colonies tab; the plate image updates as you go. See Finding colonies.
4. Choose the colonies¶
Every red colony will be scanned. To scan only some:
- pick the Lasso tool and draw around the colonies you want - colonies outside the lasso turn green;
- or pick the Hand tool and click colonies to switch them on and off;
- or set Max Colony Number on the Control tab.
See Choosing colonies.
5. Scan¶
Click Scan Colonies. The laser comes on and the stage brings each colony into the beam in turn, on the shortest route. For each one BacBuster centres the colony, sets the exposure and takes its scatter image, which appears on the left as a thumbnail. The button shows a stop sign while it works: click it to stop early. The progress bar at the bottom left shows how far it has got.
When it is done, the laser goes off and the thumbnails are ticked or unticked according to their quality.
6. Review¶
Click thumbnails to see each scatter image large. Untick the ones you do not want - off-centre, overlapping, blurred - and click Click to Purge Unchecked Colonies to remove them. Raising My Goodness Level on the Control tab unticks more of them at once. See Reviewing scatter images.
7. Save¶
Click Save plate. The Plate Info dialog asks for the plate's name, user, experiment, organism and the rest - filled in already if you dragged a planned plate. Then:
- Save To Database saves it to the database on the toolbar;
- Save writes a
.pltfile; - Export writes a folder with the plate image and one image per colony, which Baclan reads.
See Plate information and saving.
8. Next plate¶
Click Load / Eject, swap the dish and start again from step 3. Plate Image clears the previous plate - if it was not saved, BacBuster asks first.
One click for plates in a row
With a Cytomat incubator BacBuster does steps 2 to 7 by itself for every plate in it, as often as you like. See The Cytomat incubator.