Troubleshooting

The Log tab, and the day's file in C:\BacBuster\log\, say what BacBuster did and what went wrong. Look there first.

Starting

Message What to do
Device #1 is not connected! The motor controller did not answer. Check that the instrument is on and its cable plugged in, and that Ports on the Staff Only tab names the right port; restart
Device #1: No Plate Camera, No Scatter Camera The camera was not found. Check it in pylon Viewer; if it shows there, choose it when BacBuster asks, or use Camera Settings
BacBuster is already running in the background and does not respond An earlier BacBuster got stuck while closing and holds the cameras. Answer Yes to end it
Can not find out a device model!!! No device configuration was chosen - see Devices and cameras

Scanning

Problem What to do
Error taking Plate Image! The plate camera gave no picture. Check it live under the live plate view
Too many or too few colonies Adjust the Find Colonies settings - see Finding colonies
Plate image and colonies do not line up with the laser Calibrate again - see Calibration
Scatter patterns off-centre Raise Max Steps, check Use Laser Offset, or calibrate again
Warning! Exposure times are variable!!! The exposure changed during a scan without auto exposure: scan the plate again
Plate light or laser goes off by itself LEDsaver and LaserSaver on the Staff Only tab

Saving

Message What to do
Plate "..." is not saved. Would you like to save it? Yes to save, No to discard, Cancel to go back
Not connected to Database Click the database label and choose a database that works, or save to a file
Plate Name contains forbidden characters or is empty Give the plate a name without \ / : * ? " < > |
Duration of Incubation Format is wrong, Wrong time format Correct the dates and hours in the Plate Info dialog

The red warning button

Hover over it to see why it is there, and click it to dismiss it:

Hint Meaning
Default Database Is NOT Set or Connection is lost No database at start-up - see Connecting to a database
Plate was not saved! An unattended run could not save a plate; a Cytomat run left it in C:\BacBuster\Emergency\
Cytomat Error The Cytomat reported an error - see its window

Cytomat

Problem What to do
Runs are due but nothing happens A plate is still marked as out. Make sure none is in the instrument, then click Plate Out
The Error light is red See the Cytomat's own display; Reset Error on the Advanced tab clears it
No plates offered for a place Click the window's database label and choose the database the plans are in