Test Drive

The Service > Test Drive tab scans the dish that is in the instrument again and again, at fixed intervals, between two dates. It is a time-lapse of one plate - to follow colonies as they grow - and a way to test an instrument over hours or days.

Setting it up

Field Meaning
First run Date&Time When the first scan is due
End Date&Time No scans after this
Interval (MIN) Minutes between scans, counted from the first one
Save Colonies Ticked: every scan takes the plate image and scans the colonies. Unticked: the plate image only
Save Plate Saves every scan to the database on the toolbar
Set Default Plate Info Opens the Plate Info dialog for the description the saved scans get; Set Plate Info keeps it

Without a description, saved scans are named Test1, Test2 and so on. The incubation time of each is counted from the plating date of the description.

Click Start. TestCount counts the scans done; Stop ends the run early. The run also ends by itself at the end date, with the message Done. The settings and the state of a run are kept, so a run survives a restart of BacBuster.

When a scan cannot be saved to the database, the red warning button appears and the e-mail alert goes out.

Following the same colonies

Tick UseCache on the Cache tab so that every scan follows the colonies of the first one as they grow - see Scanning colonies.

The other buttons

Button Does
Test Plate Image Loads a picture from a file as the plate image and finds colonies on it
Take Stack of Images Takes 128 pictures with the scatter camera as fast as it can, and writes their timing and noise statistics to files - for testing cameras
AutoExposureTest Steps the scatter camera's exposure from shortest to longest and plots what every auto-exposure measure gives - for choosing the auto-exposure algorithm

Panorama

On instruments with a microscope camera, Panorama turns the test drive into a scan of a grid of microscope pictures: X and Y Start, Step and End (mm) set the grid, Set Microscop Start Position and Go To Microscop Start Position its starting point, and the pictures go to a folder chosen at Start, shown as Panorama Directory.