Devices and cameras

BacBuster drives several kinds of BARDOT instrument. Which one it drives, under what name, and which cameras are which, is set once per computer.

Device configurations

An instrument is described by a device configuration: a .cfg file in the program folder whose [Gizmo] section gives the model and the name of the instrument:

Model Instrument
PUCL The BARDOT with an X-Y stage on a Velmex motor controller
MINI1, MINI2 The Mini BARDOT, whose stage turns the dish and slides it
HETTICH A pair of instruments, A and B, sharing one plate loader. The name of the second is DeviceName2
NONE No instrument

Change Device on the Service > Staff Only tab opens the Gizmo Selector, which lists the configurations found; choose one, click OK and restart BacBuster. None starts it without an instrument. The choice replaces the [Gizmo] section of BacBuster.ini - which also holds the camera assignments, so BacBuster asks which camera is which at the next start.

The name in the configuration names the instrument's own files in the program folder:

File Holds
<name>.ini The instrument's settings: positions, exposures, ports, colony-finding settings, window layout
<name>.tab The plate camera's distortion map, from calibration
<name>.bgr The plate camera's background, from calibration
<name>.cpt The calibration's vectors

So moving an instrument to another computer means copying these files with it; and a computer that runs two instruments keeps them apart by name.

Cameras

Each instrument has a plate camera, which photographs the dish, and a scatter camera (or laser camera), which photographs the scattered light; some have a microscope camera too. BacBuster tells them apart by serial number, kept in BacBuster.ini. When a camera it expects is not connected, BacBuster asks at start-up which of the connected cameras to use instead.

Control (Service > Staff Only) Does
Swap Cameras Swaps the plate and scatter cameras, for when they were assigned the wrong way round. Restart BacBuster afterwards
Camera Settings Opens the Camera Selector: pick a camera to see it live, set its Duty (which camera of which device it is), and Flip Vertical, Flip Horizontal or Turn 90 degrees its picture

Replace Laser camera with Microscope one on the Big Bardot tab scans with the microscope camera in place of the scatter camera.

Basler cameras need the pylon Software Suite installed; PixeLINK cameras need theirs. A camera that pylon Viewer cannot see, BacBuster cannot see either.

Serial ports

The Ports box on the Staff Only tab chooses the serial ports of the Relay (laser and plate light), the Motor controller, the Heater and the Cytomat. Choose the ports Windows gave the instrument's adapters - Device Manager lists them - and restart BacBuster.

Lights and safety

Control (Staff Only) Does
LEDsaver Switches the plate light off after this many minutes without activity. 0 never
LaserSaver Switches the laser off after this many minutes without activity. 0 never
External Laser The instrument's laser is outside it and always on
Eclipse button Moves the stage to where it blocks an external laser
Set Current Plate Holder Position as Laser Eclipse Position Sets where that is
Park button Moves the stage to its parking position
Fan, Coil buttons Switch the heater's fan and coil by hand
100% motor power The motors run at 70 % power; tick this if they skip steps

Two instruments

A HETTICH configuration runs two instruments, A and B, from one BacBuster: each gets its own main window, and a Remote window shows the two side by side. The plate of instrument A is marked yellow on the Project tab, that of B aqua.

No instrument

Tick NoGizmo on the Staff Only tab and restart, or choose None in the Gizmo Selector, to run BacBuster without an instrument - see No-device mode.