The plate image¶
The plate image is the photograph of the whole dish that colonies are found on. It is the first thing BacBuster takes of every plate, and it is saved with the plate.
Taking it¶
Click Plate Image on the toolbar. BacBuster:
- asks whether to save the current plate, if it has scanned colonies that were not saved;
- switches the laser off and the plate light on;
- moves the stage to the plate position, and takes the picture with the plate camera;
- switches the plate light off;
- removes the uneven background and finds the colonies - see Finding colonies;
- shows the picture below the live views, with the colonies found coloured red.
The button does nothing while colonies are being scanned, and nothing in no-device mode. Error taking Plate Image! means the plate camera did not deliver a picture.
Watch it live first
The camera button under the live plate view shows the plate camera live, and the light button next to it switches the plate light on, so that you can check the dish is in place and the exposure is right before taking the picture. The light goes off by itself after a few minutes - see LEDsaver on the Service tab.
The area of the dish¶
Only colonies inside the dish's area are looked for. The Control tab sets it, in pixels of the plate camera:
| Setting | Meaning |
|---|---|
| Round Plate | Ticked for round dishes, unticked for rectangular ones |
| Radius | Radius of the area searched on a round dish |
| Plate Width, Plate Height | Size of the area searched on a rectangular dish |
Calibration sets them for the instrument's standard dish; change them for other dishes, and keep them in a profile.
Exposure of the plate image¶
The slider under the live plate view sets the plate camera's exposure; double-clicking the live view returns it to the default that calibration set. Two more settings on the Control tab help with faint colonies:
| Setting | Meaning |
|---|---|
| Plate image reps | Number of pictures added together into one plate image. 1 takes a single picture |
| Gain | With more than one picture: the contrast curve applied to the sum. Positive values brighten the faint parts, negative values darken them; 0 or 1 leaves it nearly linear |
Image options¶
On the Options tab:
| Option | Effect |
|---|---|
| Raw Plate Image | Keeps the plate image as the camera took it: no distortion correction, no normalising, no inverting and no mask around the dish |
| Take Raw Image | Takes a plate image where the stage is now, without moving it and without looking for colonies |
| Positive Image | Leaves the plate image positive. Normally it is inverted, so that colonies come out bright on a dark background; the new calibration plate needs it positive |
| Colorify | Shows scatter images in false colour, which makes faint detail easier to see; it does not change what is saved |
Warning
Turn Colorify, Raw Plate Image and Positive Image off before a Cytomat run; BacBuster reminds you when you start one with any of them on.
Printing it¶
Right-click the plate image and choose Print to print it at its real size, so that a print laid under the dish matches the colonies.
Saving the picture alone¶
Save Plate Image and Load Plate Image on the Service > Staff Only tab save the plate camera's picture as an image file, and load one back to find colonies on it - useful for trying out the colony-finding settings on a known picture.