Baclan identifies bacterial colonies from their scatter patterns. A BARDOT instrument, driven by the BacBuster program, finds the colonies on a Petri dish, shines a laser through each of them and photographs the pattern of light scattered forward. Baclan takes those images, turns each one into a vector of numeric features, trains a support vector machine on colonies of known organisms, and then says to which class each colony of a new plate belongs, and how probable that is. Its title bar calls it BacteriumAnalyzer.

This manual describes Baclan version 12.22 (Windows 64-bit).

The main window after classifying a plate

What Baclan does

Keeps colonies in plates and plates in classes. A plate is the picture of a dish with the scatter images of its colonies. A class is a named list of plates. Plates come from folders, from .plt files or from a database.

Turns images into numbers. Filters prepare each image, then extractors calculate Zernike moments, Haralick textures and other features from it. The list of filters and features, with their settings, is up to you.

Trains and checks a classifier. A grid search finds the best SVM parameters, and a cross-validation matrix shows how well the classes can be told apart.

Classifies new plates. Every colony of a test plate gets a class and a probability. The plate image shows them as coloured circles, and a report goes to the plate's folder.

Looks at data without a model. Colonies can be clustered, and mapped in two dimensions with PCA, tSNE, UMAP or an autoencoder.

Keeps everything in a database, if you want. Plates, classes and models can live on a MySQL or MariaDB server instead of in files, where they can be searched and shared.

Where to start

Words used in this manual

Word Meaning
Colony One scatter image, with the position and size of the colony on the plate. Also called a sample
Plate The image of a dish and its colonies
Class A named list of plates, all of whose ticked colonies are taken to be one kind of organism
Model The classes, the filters and features chosen, and the trained SVM, saved together
Training Set The classes of the model
Test Set The plates to be classified

Conventions used in this manual

Convention Meaning
Shift +click Hold the key and click
Bold A button, tab, menu item or check box you can see on screen
Monospace A file name, folder name or literal value

Asking for help inside Baclan

F1 opens the page of this manual about whatever is under the mouse. The Help button on the toolbar opens its front page. Toolbar buttons that show only an icon say what they do when you hover over them.