PCA, tSNE, UMAP and the autoencoder

The PCA tSNE UMAP etc tab maps the colonies of the Test Set onto a plane, one point per colony, so that colonies with similar features fall near each other. It uses the filters and features chosen on the Features tab, each feature normalised, and needs at least four colonies. No model is involved, and none is changed.

The PCA tSNE UMAP etc tab

Button Method
natSNE tSNE, by FIt-SNE. A plate with few colonies gets a smaller perplexity by itself
CAE A contractive autoencoder, trained on the colonies for the number of Epochs beside it
PCA The first two principal components
UMAP UMAP, with the settings of Set UmapPars
Set UmapPars The UMAP Parameters dialog: the parameters of umap-learn, by their names. Set Defaults brings back the defaults. The metric can be euclidean, manhattan, cosine or correlation
Overview Thumbnails of the colonies, one row per class
VAE Decoder Opens the Scatter Generator
Point Size The size of the points

tSNE, UMAP and the autoencoder run in the background, one at a time. The button reads Stop while its method runs, and the status bar shows how far it has got. Press it to stop. The chart clears when a method starts and is drawn when it ends; if the plates changed meanwhile, Baclan says to run it again.

The chart's title names the method, with a number: how far the distances on the map differ from the distances between the features. Smaller is better, for comparing methods on the same colonies.

Each point has the colour of its plate. Double-click a plate in the Test Set tree to change it, for the next map drawn.

Picking colonies on the map

Click a point to make its colony active: its plate image, thumbnail and table row follow.

To gate a group of points:

  1. Hold Shift and click on the chart where the outline should start.
  2. Click for each further corner.
  3. Right-click to finish.

The colonies inside become class 1, those in a second outline class 2, and so on; all the others are class 0. Their circles on the plate image change colour accordingly.

To Do
Move an outline Drag it. The classes follow when you let go
Move one corner Drag the corner
Rotate an outline Drag it with the right button
Delete an outline Ctrl + click it. The classes follow with the next change
See its colonies Hold the mouse over it: a strip of their thumbnails opens

These classes are for looking: they stay with the Test Set and do not change the model.

The Scatter Generator

VAE Decoder opens a window that draws scatter images from a decoder network, with a slider for each of its inputs. Choose the decoder from the list, from the .json files in %APPDATA%\Baclan\Keras\Decoders. The list is empty unless you have put decoders there.