Cross-validation¶
The CV Matrix tab shows how well the model tells its classes apart. In cross-validation the training colonies are split into parts; each part in turn is classified by an SVM trained on all the others. So every colony is classified by a model that has not seen it.

Baclan runs this at the end of every training. To run it again with other settings, change them and press Recalc.
| Control | Does |
|---|---|
| % | Ticked, each row is in per cent of that row's colonies. Unticked, it shows numbers of colonies |
| Export | Opens a menu: Features writes the features of all the training colonies to a text file, Crossvalidation Table the matrix as shown |
| n-Fold | Into how many parts the colonies are split |
| Reps | How many times the whole thing is repeated, with a new split each time. The numbers are the averages |
| Recalc | Runs it again, with a progress window. Abort there leaves the old matrix |
Reading the matrix¶
Each row is a class as you defined it. Each column is what the cross-validation called those colonies. So the diagonal is what it got right, and in the row of class A a large number under class B means that colonies of A are being taken for B.
| What you see | What to look at |
|---|---|
| A low diagonal for one class | Its plates: mistaken colonies, artefacts left ticked, a different medium or incubation time |
| Two classes mixed with each other | Whether the organisms can be told apart at all. Other features, or more plates |
| All close to 100 % with few colonies | Whether colonies of one plate are too alike. Test it on plates it has not seen |
Note
This measures how well the model separates the colonies it was trained on. Plates grown on another day are the harder test: put them in the Test Set.
The matrix and its n-Fold and Reps are saved with the model. Recalculating marks the model as changed.