Save Plate

Save Plate on the top toolbar opens the Plate Info dialog for the active plate.

Field Meaning
Plate Name The plate's name
User, Experiment, Lab, Media, Source Who, what for and on what. Each list offers what is in the database, when one is connected
Catalog Number The organism, by its catalogue number. New and Edit open the Organism dialog: genus, species, subspecies, serotype, pathotype, resistance, pathogenicity. They need a database
Device ID, Robot ID Which instrument and robot
Barcode The plate's barcode
Incubation date&time When it went into the incubator
Duration of Incubation Hours and minutes
Temperature The incubator's temperature, in °C
Originated from Click to choose the colony this plate was grown from
Date and Time When the image was taken. It cannot be edited

Then choose where the plate goes:

Button Does
Save Writes one .plt file. A high-dynamic-range plate writes a second one, <name>_HDR.plt
Export Writes a plate folder in a folder you choose, named after the plate
Save To Database Saves the plate in the connected database. It is available only when Baclan can connect to the database named on the toolbar
Close Closes the dialog. The plate stays as it was

If the plate is in the database already, a Warning asks whether to Replace Entire Plate or to Replace Plate Info only, or to Cancel.

Export as PNG file list

Export as PNG file list, the third button of the Test Set toolbar, asks for a folder and writes the active plate into a subfolder named after the plate and its incubation time.

What goes in, with <name> the name of that subfolder:

File Holds
<name>_N001.png, ... The scatter images
<name>_N001_HDR.png, ... Their high-dynamic-range versions, if the plate has them
<name>_N001.csv, ... How BacBuster centred the colony, if the plate records it
SCV.txt, ColonyXY.csv The table of colonies
METADATA.txt The plate's details
AIM.png The image of the plate

An existing folder is emptied first

If the folder exists and holds images or tables, Baclan asks Would you like to delete the directory with all scatter files?. Yes deletes every .bmp, .png, .csv and .txt file in it; Cancel stops the export.

Exporting many plates

The Database Control Panel exports whole selections from the database, also smaller images and filtered ones, for training other classifiers.

Save Class

Save Class on the top toolbar saves the selected class to the database, or in Test Mode all the test plates as one class under a name you give. It works only with a database: with the switch on files it says This feature works with a database only!!!.