Bead gates¶
Bead gates separate the bead populations. Each gate becomes one analyte once it has been identified. All of this happens on the Scattered tab.
The Scattered tab¶
| Part | Content |
|---|---|
| BEADS list | The analytes of the current assay with their bead numbers |
| ASSAY | The current assay. Click the name to choose another |
| REPORTER | The parameter whose intensity is converted to concentration |
| Chart | The events on the two bead classification parameters, coloured by gate |
| Use Pregate | Apply the noise gate |
| Scroll bars along the axes | Fine compensation - see below |
| Auto Gate, Semi Auto, Rename, Save Layout | Gate commands |
Click an axis of the chart to choose its parameter; click the corner between the axes to swap them. Double-click the chart to open the chart editor.
Finding the gates automatically¶
Choose the assay and the reporter, set the axes to the two classification parameters, and press Auto Gate. Mplex clears the existing gates and searches the visible part of the chart for bead populations, creating one gate for each.
Zoom the chart to the bead cloud first if there is debris elsewhere; only the visible region is searched.
Semi-automatic gating¶
Semi Auto runs the same search, but first opens the Semiauto Gating window so you can guide it. The window shows the bead cloud as a density map, with the regions found so far outlined in red.
| Control | Effect |
|---|---|
| Threshold slider | The density above which a point belongs to a bead region. Raise it to split regions that run together; lower it to take in the faint edges of a region |
| X Filter, Y Filter | How much the map is smoothed along each axis before the regions are found |
| Drag on the map | Draw a separating line - use it to cut apart two populations that touch |
| greedy | Enlarge the gates more generously when they are created |
| Create Gates | Create the gates from the regions shown |
| Close | Cancel without changing the gates |
Use it when Auto Gate merges neighbouring beads or misses dim ones.
Note
Gating needs an assay and at least five standard wells. Without them Mplex asks you to choose an assay, or to mark calibration wells on the plate.
Drawing and deleting gates by hand¶
| Action | Result |
|---|---|
| Shift +click on the chart | Start a new gate; click to add vertices, click the first vertex to close it |
| Drag a vertex or the gate | Reshape or move it |
| Shift +right-click inside a gate | Delete the gate |
A gate you draw or name yourself is user-defined.
Identifying the beads¶
A gate becomes an analyte when it knows its bead:
- By hand - drag a row from the BEADS list onto a gate. The gate takes the bead number and analyte name. If another gate had that bead, it loses it.
- From the layout - if the assay has a saved bead layout, Rename names every gate after the nearest saved bead position.
When all gates are named, press Save Layout to store their positions in the assay for next time.
The gate list¶
The ALL GATES list in the results area has one row per bead gate:
| Column | Click to |
|---|---|
| First (delete) | Delete the gate |
| Name columns | Select the gate - the chart and plate map then show only its events |
| Colour | Change the gate's colour |
Tick ALL GATES to show all gates together again.
Histograms¶
The Histogram tab shows the reporter histogram of the selected gate, or of all gates when ALL GATES is ticked. Click the X axis to choose another parameter; double-click for the chart editor.
Compensation on the chart¶
The scroll bars along the chart's axes shift one parameter against the other to correct spillover between the reporter and a classification parameter. The current value is shown while you drag. The full spillover matrix is on the Parameters tab.