A complete analysis¶
This page takes one plate from raw FCS files to a table of concentrations. Each step links to the page that covers it in detail.
Before you start¶
- The plate's FCS files are in one folder, one file per well.
- The kit is in the assay database: its analytes, their bead numbers and the top standard concentration of each.
- You know the dilution series of the standards: the dilution factor between steps, and the number of steps.
1. Open the plate¶
Click the open-plate button (the first on the toolbar) and choose the plate folder. If the plate has been analysed before, its legend, excluded wells and default gates are loaded with it. See Opening a plate.
Tip
For a very large plate, press Preview before opening. Set-up is much faster on a fraction of the events; turn preview off and reopen for the final numbers.
2. Mark the standards¶
Open the Legend tab. Type c followed by the dilution step into the first field of
every standard well - c1 for the most dilute standard up to, say, c7 for the top
standard, and c0 for the blank. Enter the Dilution Factor and Number of
Dilutions and press Apply. See The plate legend and standards.
3. Choose the assay and the reporter¶
On the Scattered tab, click the assay name under ASSAY and choose the kit. Under REPORTER choose the parameter that measures the analyte (PE, for most kits).
4. Remove noise¶
On the Noise Gate tab, draw a gate around the singlet beads and tick Use Pregate on the Scattered tab. See Noise gates.
5. Gate the beads¶
Choose the two bead classification parameters as the axes of the Scattered chart and press Auto Gate. Mplex finds one gate per bead population. Check them, adjust any that are wrong, and draw missing ones with Shift +click.
Identify the beads: drag each name from the bead list onto its gate. If the assay already has a saved bead layout this happens by itself - otherwise press Save Layout once the gates are named, and the next plate will be identified automatically. See Bead gates.
6. Check the calibration¶
The calibration chart shows the standards and the fitted curve for every analyte ticked in Calibrated Analytes; the list shows R² for each. Exclude a failed standard well by right-clicking it on the plate map. See Calibration curves.
7. Export¶
Export CSV writes concentrations, intensities, spread and counts of every well for the ticked analytes. Export Calibrations writes the fitted curves. See Exporting results.
8. Save¶
Set As Default stores the gates with the plate, so they reload whenever the plate is
opened. Save project writes them to a .pap file you can load on other plates. See
Projects and defaults.