Noise gates¶
A noise gate keeps only the events worth analysing - single beads rather than debris, doublets or aggregates. Mplex lets you chain several 2-D gates; an event has to pass all of them.
Making a noise gate¶
The Noise Gate tab shows one page per gate, plus a + page.
- Click + to add a gate page.
- Choose the X and Y parameters by clicking the chart's axes - typically forward and side scatter first.
- Hold Shift and click on the chart to start a gate, then click around the population to add vertices. The gate closes when you click its first vertex.
- Add further pages for further conditions, e.g. a doublet gate on area versus height.
The chart menu also offers Native Coordinates, Unzoom and Close Page. The + and - buttons change the dot size; double-click the chart to open the chart editor.
To delete a gate, hold Shift and right-click inside it.
All gates share one colour, chosen in the colour box of any page.
Using the noise gate¶
Tick Use Pregate on the Scattered tab. Only events inside every noise gate then reach the bead gates, the heat maps and the calibration.
The noise gate is saved with the project and with the plate defaults.
Saving the gated events¶
The Save gated toolbar button writes a new plate that contains only the events
inside the noise gate (all events if Use Pregate is off). It is created as a new folder
Scion1, Scion2, ... inside the plate folder, with the plate's legend and INFO folder
copied, and opened straight away.
This is useful for large plates: gate once, save, and work on the much smaller plate from then on.