Exporting results

All exports are comma-separated text that opens in any spreadsheet program. Plate-shaped tables have one row per plate row and one column per plate column.

Export CSV

Export CSV under Calibrated Analytes writes one file for the whole plate. For every analyte of the assay it contains nine plate-shaped tables, grouped by quantity:

Section Content
Median Concentrations Concentration from the median reporter intensity
Concentrations Concentration from the mean intensity
Concentration Median Deviation Median absolute deviation of concentration
Concentration Std Deviations Standard deviation of concentration
Bead Counts Number of beads of this analyte in the well
Median Intensities Median reporter intensity
Intensities Mean reporter intensity
Intensity Median Deviation Median absolute deviation of intensity
Intensity Std Deviations Standard deviation of intensity

Each section starts with its title; each analyte within it starts with the analyte's name. An analyte of the assay that has no gate is written as a table of zeros, so every file of the same assay has the same layout.

Check the Bead Counts before trusting a value: a well with only a handful of beads gives an unreliable median.

Export Calibrations

Export Calibrations writes the fitted curves. For every analyte:

IL-6,B4
Maximum asymptote = 21834.5
Minimum asymptote = 12.7
Inflection point = 1893.2
Hill's slope = 1.08
Asymmetry = 0.94
Expected C,Calibrated C,Intensity
10000,9731.2,20113
...

The parameters are those of the fitted model: the asymptotes are log10 intensities and the inflection point is on the ln-concentration scale. Asymmetry appears only for the 5-parameter model. The table lists, for every standard, the concentration it should have (Expected C), the concentration the curve gives for its intensity (Calibrated C) and the intensity itself - a direct check of the fit.

To CSV in the legend editor

To CSV in the plate legend editor writes one long table with a row per analyte per well:

Analyte, Sample, <value>, CV%, Count

where Sample is the well's legend and value is the quantity chosen in Heat Map Values. Each well's rows are preceded by the well name.

Export All

Export All on the Options tab saves the whole plate - every event - as a text file.

Pictures

Copy above the plate map and above the multi heat map copies them to the clipboard. Charts are copied or exported from their chart editor (double-click the chart).

Description of the files

A text description of the export files, with the formula of the curve, is installed in %APPDATA%\Mplex\Docs\readme.txt.