Heat maps

The Plate Map tab

The plate map colours every well by how many of its events fall in the current gate - the selected bead gate, or all bead gates when ALL GATES is ticked.

Control Effect
Heat Map Values Percent of the well's events, or their Count
Show Selected Wells Show only the selected wells in the charts
Colour scale Click it to switch automatic scaling of the colour range on or off
Export Write the plate map to a file
Copy Copy the plate map to the clipboard

Point at a well to see, in the status bar, its name, its value and the number of events that passed the noise gate, together with the well's file name.

Click a well to select it, drag to select several; the charts then show only the selected wells. Right-click a well to exclude it.

The multi heat map

The multi heat map in the results area shows several analytes at once. Every well is divided into as many segments as there are ticked analytes in Calibrated Analytes, each segment coloured by that analyte's value in that well.

Heat Map Values above it chooses the value:

  • Mean Concentration
  • Mean Intensity
  • Median Concentration
  • Median Intensity

Point at a segment and its analyte is selected in the list. The colour scale beside the map shows the value range. Copy copies the map as a picture.

Tick a single analyte (or tick Exclusive) to see one analyte across the whole plate.