Troubleshooting¶
"Cannot open Plate"¶
The folder holds no FCS files Mplex can read, or they cannot be mapped onto wells. Check the folder, and try Rename on a plate that did open to give the files well-based names.
"Select an assay, pls."¶
Choose an assay under ASSAY on the Scattered tab, or on the Assay tab, before gating.
"Select the MPLEX parameter"¶
Choose the reporter under REPORTER on the Scattered tab.
"Not enough calibration points"¶
Fewer than five standard wells were found. Mark the standards in the
plate legend - c followed by the dilution step - and press
Apply.
A calibration curve is flat or has a low R²¶
- The bead gate may be wrong or mis-named - check it on the Scattered tab.
- The standards may be numbered the wrong way round -
c1is the most dilute. - The Dilution Factor or Number of Dilutions may not match the plate.
- One standard well may be bad - exclude it with a right-click on the plate map.
"Bead name mismatch" or "Gate number mismatch" on Save Layout¶
Save Layout needs exactly one named gate for every analyte of the assay. Delete extra gates, draw missing ones and name them all first.
Everything is slow¶
Press Preview and reopen the plate while setting up; or save a noise-gated copy with Save gated and work on that.
The gates are gone when I reopen the plate¶
They were not stored with the plate. Press Set As Default after finishing a plate.
Resetting Mplex¶
Close Mplex and delete %APPDATA%\Mplex\mplex.ini. Your assays, layouts and plates are
kept.