Skip to content

Calibration curves

For every analyte Mplex fits a curve of reporter intensity against concentration through the standard wells, and reads the concentration of every other well off that curve.

The model

Mplex fits a Hill (logistic) curve to the logarithm of the reporter intensity S against the natural logarithm of the concentration x:

4-parameter curve

\[ \log_{10} S = d + \frac{a - d}{1 + \left(\dfrac{u}{c}\right)^{b}}, \qquad u = \ln x + s \]

5-parameter curve - adds an asymmetry exponent g

\[ \log_{10} S = d + \frac{a - d}{\left(1 + \left(\dfrac{u}{c}\right)^{b}\right)^{g}} \]
Symbol Name Meaning
d Maximum asymptote log10 intensity at very high concentration
a Minimum asymptote log10 intensity at the low end of the curve
c Inflection point The position of the curve's mid-point, on the u scale
b Hill's slope How steep the curve is
g Asymmetry 5-parameter only: how far the curve departs from symmetry about c
s Shift A whole number added when some standards are below 1, so that u is positive; 0 otherwise

Working on logarithms keeps a standard series spanning several decades evenly weighted. The blank (c0) has no logarithm and does not take part in the fit.

Choose the model in Calibration Curve on the Options tab. The 5-parameter model usually fits wide standard ranges better; the 4-parameter model is more stable with few standards.

Settings

All on the Options tab:

Setting Effect
Calibration Curve 4-parameter or 5-parameter
Use Median Intensity For Calibration Curves Use the median rather than the mean reporter intensity of each standard well. The median is less affected by stray events
Optimisation Algorithm The fitting method: Marquardt, Simplex, BFGS or Simulated Annealing. BFGS is the default and works well in most cases

Every change refits all curves at once.

Reading the curves

The calibration chart shows, for every analyte ticked in Calibrated Analytes, the standards as points (observed) and the fitted curve as a line (expected), in the analyte's colour, on log-log axes. Double-click the chart for the chart editor.

Calibrated Analytes lists every analyte with the R Sqr (R²) of its fit. Values close to 1 mean the curve passes through the standards; a low value points to a bad standard, a wrong dilution series or a mis-identified bead.

Control Effect
Tick box Show this analyte in the chart and the multi heat map
All Tick every analyte
Exclusive Allow only one ticked analyte at a time - clicking an analyte shows it alone

Improving a curve

  • Exclude a bad standard - right-click the well on the plate map. The curve is refitted without it.
  • Check the legend - a standard with the wrong c number bends the curve.
  • Check the dilution series - the Dilution Factor and Number of Dilutions in the legend editor must match the plate.
  • Check the bead gate - a gate that catches part of a neighbouring bead gives a flat or noisy curve.

Concentrations outside the range of the standards are extrapolated and should be treated with caution.