BacBuster

BacBuster runs a BARDOT instrument, the bacterial colony scatterometer. It photographs a Petri dish, finds the colonies on it, then moves the dish so that a laser beam passes through each colony in turn, and records the pattern of light the colony scatters forward: its scatter image. The plate image, the colonies' positions and their scatter images are saved together as a plate, in a file or in a database, from which Baclan classifies them.

This manual describes BacBuster version 12.33 (Windows 64-bit).

The main window with a scanned plate

What BacBuster does

Takes the plate image. The plate camera photographs the dish under the plate light, and BacBuster finds the colonies on that image by their brightness, size and roundness.

Lets you choose the colonies. All colonies found are scanned unless you leave some out: draw around the ones you want, click colonies in or out, or cap their number.

Scans each colony. The stage brings every chosen colony into the laser beam, centres it, sets the exposure and takes its scatter image - in low or high dynamic range, or both.

Grades the images. A trained network scores every scatter image; the poor ones come unticked, and you can purge them before saving.

Saves the plate. With its name, barcode, organism, medium and incubation time, to a MySQL or MariaDB database, to a .plt file or to a folder of images.

Runs unattended. With a Cytomat incubator BacBuster takes plates out at the times you set, scans them and puts them back, for as long as the experiment lasts.

Where to start

Words used in this manual

Word Meaning
Plate A dish: its plate image, the colonies found on it and their scatter images, with its plate information
Plate image The photograph of the whole dish, taken by the plate camera
Colony A spot found on the plate image. Scanned colonies are also called samples
Scatter image The pattern of laser light scattered forward by one colony, taken by the scatter (laser) camera
Stage The motorised holder that moves the dish
Device One BARDOT instrument, with its own settings and calibration files
Plate information Name, barcode, user, experiment, organism, medium, source, lab and incubation of a plate

Conventions used in this manual

Convention Meaning
Shift +click Hold the key and click
Bold A button, tab, check box or field you can see on screen
Monospace A file name, folder name or literal value

Asking for help inside BacBuster

F1 opens the page of this manual about whatever is under the mouse. The Help button on the toolbar opens its front page.

Safety

The BARDOT laser is on while colonies are scanned and while the laser button is down. Do not look into the beam.